mis alignment of detail scans

Post by shollinger70 »

I have found that my detail scans from the same scanworld are not aligning with the original scan. They seem to be about a half inch lower. We use a P30. The scanner was level and was at grade.
See attachment.
You do not have the required permissions to view the files attached to this post.
tbwester
V.I.P Member
V.I.P Member
Posts: 904
Joined: Fri Apr 20, 2012 1:49 pm
11
Full Name: Thad Wester
Company Details: Clarity Scanning
Company Position Title: President
Country: USA
Linkedin Profile: Yes
Location: SC
Has thanked: 1 time
Been thanked: 2 times

Re: mis alignment of detail scans

Post by tbwester »

probably moved the scanner when you boxed out the area with your finger or Leica plastic pen. It is possible to move it enough to make it misaligned but not enough that the scanner would notice it is no longer level -- in my experience.
Hans Tuexsen
I have made <0 posts
I have made <0 posts
Posts: 6
Joined: Wed Aug 03, 2016 2:01 pm
7
Full Name: Hans-Herbert Tuexsen
Company Details: Leica Geosystems
Company Position Title: Product Manager TLS
Country: Switzerland
Linkedin Profile: Yes
Been thanked: 1 time

Re: mis alignment of detail scans

Post by Hans Tuexsen »

When the scanner is setup on a solid tripod it will not move when you define a detail scan area with the stylus or your finger. When the scanner setup was stable, which I assume it was, then a offset between the overview scan and a detail scan (child scan) could indicate a problem with the calibration of the scanner. To check this do a 2-face target scan and check the delta values in the Delta tab of the Target Information panel.
You do not have the required permissions to view the files attached to this post.
shollinger70
I have made 10-20 posts
I have made 10-20 posts
Posts: 15
Joined: Mon Jan 09, 2017 1:26 pm
7
Full Name: Stephen Hollinger
Company Details: HGA
Company Position Title: 3d tech manager
Country: USA
Linkedin Profile: Yes
Has thanked: 1 time

Re: mis alignment of detail scans

Post by shollinger70 »

I just performed the 2 face check, results were fine. So looks like I just need to calibrate the tilt comp. Still going to send the unit in for year cal/maintenance with in the next couple of weeks.

Thanks for your input and help with this.
User avatar
Phil Marsh
Administrator
Administrator
Posts: 2490
Joined: Sun Aug 12, 2007 2:28 pm
16
Full Name: Phil Marsh
Company Details: Laser Scanning Forum Ltd
Company Position Title: Director
Country: UK
Skype Name: phil.h.marsh
Linkedin Profile: Yes
Location: Barnsley, UK
Has thanked: 117 times
Been thanked: 36 times
Contact:

Re: mis alignment of detail scans

Post by Phil Marsh »

It sounds like the two faces of the unit are out of alignment. I would haver the unit calibrated to make sure its all working as it should be.